Unreliable, basically /not usable SPO2 sensor.

Myself and 2 friends of mine have Foreunner 245. According the Garmin Forum replies all of us suffer of night apnea :)

A bad reading during the night is possible due to the unknown position of the wrist/watch, however even running manual tests all of us get unexpected values as 94% or so.

Therefore we collect 3 "regular" SPO2 meters and run 30 comparison tests:

Regular SPO2 tester:

- all 3 of us, 30 measurements (each minute) with 3 different testers = 98% - 99% (max reading range)

Forerunner:

- all 3 watches on all 3 of us, 30 measurements (each 3 minutes slower reading, 3 watches  each) = 92% -99% max range

At our age, lower of 97% means health issues

Conclusions:

Now we ignore SPO2, even if unluckily we chose this model especially for this unreliable feature

We were hoping of a better management/calibration with a new Firmware, unfortunately after 3 years we didn't see any improvement about SPO2 performance .